(4) (a) What does the z-value tell us about the heat resistance of a microorganism?

(b) What is the relationship between the F value and the SAL value?

(c) For a steam sterilization process at 134 °C, an F0 value of 50 min is achieved. How is the achieved F0 value of 50 min to be interpreted?

(2 points)

a) Z-value = Temperature change in °C required to alter D-value by factor of 10

  • high z-value = means large increase in °C is required to achieve 1 log reduction of D-value = means that germs are more heat resistant to changes of temeprature

  • low z-value = low increase in °C is required to achieve 1 log reduction in the D-value = germs are less heat resistant

    e.g. spores = z-value 10°C & vegetative cells = 5,5 °C

b) Relationship between the F value and the SAL value

F-value = thermal death time = holding time [mins] at a temeprature T to reduce population N0 with specific D-value by desired number of poweres (log levels) to final value N

SAL = sterility assurance level = probability that germs may be alive

  • Relationship: with sufficient F-value [holding tims in mins] a sterilisation process can ensure that the SAL-value achieves the required standard

  • high F-value = high holding time = low probability of items remaining non-sterile, low probability that germs stay alive = means a better (lower) SAL-value

  • low F-Value = low holding time = high probability that sterilisation is not sufficient & germs stay aliba = high SAL-value


(c) For a steam sterilization process at 134 °C, an F0 value of 50 min is achieved. How is the achieved F0 value of 50 min to be interpreted?

F0-value [mins] = 1 min means the lethality of a method that works for 1 minute at 121.1 °C (250 °F) and a z-value of 10 °C (18 °F) or an equivalent temperature-time combination

= makes different steam sterilization comparable -> important for pharmacy

50min (F0-value) at 134°C = process achieves the same microbial lethality as the sterilisation at 121°C for 50 mins

(19) Fill and finish

(a) What dosing system is shown in the picture?

(b) Explain the principle of time-pressure dosing.

(c) Why does Annex 1 specify that crimping must be separated from filling and stoppering?

(d) Which test must be carried out on glass ampoules after fusion as a 100% check in addition to the visual inspection?

(e) What is the main advantage of using nested ready-to-fill syringes in tubs compared to using bulk syringes?

(f) How do you call the process for manufacturing infusion or parenteral nutrition bags?

(g) What is used to blow up the containers during the blow fill seal process?

a) (Dual) peristaltic pump

= single-use silicone tube -> tube get sandwiched between rollers (Dialyse-pumpe)

= benefits: single use/ high precision/ quick product change possible

(b) Explain the principle of time-pressure dosing.

Time-pressure system = gravity & viscosity dependent system

- benefit: ultra-high output/ CIP-SIP compartible

- not for: viscous products & products that are sensitive to silicone tubes

Principle:

- pressure inside vessel must stay constant -> precise monitoring of pressure

 - pinched valves are opened & then after defined dosed product valves get closed

 - temperature very important (high impact on viscosity) -> influences flow through

(c) Why does Annex 1 specify that crimping must be separated from filling and stoppering?

Crimping = aluminium cap gets placed on the stopper

• while crimping vials a lot of metal particles get generated -> non-viable particle => therefore, important that crimping is separated from filling area (physical separation)

=> to prevent particle contamination & good air extraction are required

(d) Which test must be carried out on glass ampoules after fusion as a 100% check in addition to the visual inspection?

-> closure by fusion, e.g. glass or plastic ampoules should be subject to 100% integrity testing.

(e) What is the main advantage of using nested ready-to-fill syringes in tubs compared to using bulk syringes?

bulk syringes = Washing, drying, siliconisation, sterilization necessary before filling

nested ready-to-fill syringes = just opening & filling & closure (ready to use) -> no washing & sterilization etc.

(f) How do you call the process for manufacturing infusion or parenteral nutrition bags? = form- fill- seal production

(g) What is used to blow up the containers during the blow fill seal process?

Blow/ fill/ seal system = purpose-built machine in which, in one continuous operation, containers are formed from a thermoplastics granulate, filled & sealed, all in one production line

Blow mould closes, bottom is welded -> blowing with sterilized compressed air

(16) Barrier systems (3 points)

(a) What is the difference between a passive and active RABS?

(b) Which cleanroom class must be used as the background environment for a RABS?

(c) What are the characteristics of an isolator called "open aseptic"?

(d) How do you perform the leak test of an isolator glove (in addition to the visual inspection)?

(a) What is the difference between a passive and active RABS?

• Active RABS = integral HEPA-filtered air supply (HEPA inside)

• Passive RABS = air supply by ceiling mounted HEPA-filters (outside)


Open, active RABS


Open, passive RABS


(b) Which cleanroom class must be used as the background environment for a RABS?

• Surrounding room classification should be ISO 7 minimum in operation -> surrounding area of RABS = class B


Isolator surrounding = for open isolator class C & for closed isolator class D


(c) What are the characteristics of an isolator called "open aseptic"?

Product protection isolators = “Aseptic isolator”


II) Open isolator systems = allows continuous material flow, hermetic seal for decontamination, with unidirectional air flow

 = designed for continuous or semi-continuous ingress and/or egress of materials (Ein-/ Austritt) during operations through one or more openings -> mouse holes

 = Openings must be engineered (e.g. using continuous overpressure) that no external air/ contaminant get into the isolator

-> has tiny openings (“mouse holes”) with inside overpressure to ensure that no particles get inside

-> min. grade C surrounding (because of holes)


(d) How do you perform the leak test of an isolator glove (in addition to the visual inspection)?


2) Leak testing:

 2.1) Pressure decay test = gloves get blowed up with pressure

 2.2) airflow test at constant pressure = air gets inside of gloves & measurement of volume that is needed to keep pressure constant => if tight, no additional air needed

(23) Hygienic design of plant components (4.5 points)

(a) Draw in a design improvement for the pictured container bottom.


(b) Name two criteria for the design of pipework that are not related to the properties of the material / surfaces.


(c) Sketch / draw the compression of a seal with an axial stop.


(d) In the picture you can see two possible sealings of the longitunial movement of a valve rod. Which one is better with regard to hygienic design / contamination risk? Please briefly justify your choice.


(e) Why is a simple mechanical seal not suitable for sealing agitator shafts?

a) Draw in a design improvement for the pictured container bottom.


(b) Name two criteria for the design of pipework that are not related to the properties of the material / surfaces.

Self-drainable -> have special slope (are not straight)

• Fixed connections prefered to detachable ones


Avoidance of

• dead legs (no flow of media) Ú rule: L < 3 x D

• sharp-edged profile extensions & reductions (no corners)

• large-scale flow around / flow shadow (creates dead leg)


c) Sketch / draw the compression of a seal with an axial stop


d) Which one is better with regard to hygienic design / contamination risk? Please briefly justify your choice.

Sealing of longitudinal movements

a) dynamic sealing

b) better: hyrmetic sealing = because hermetically sealed & no germs can enter


(e) Why is a simple mechanical seal not suitable for sealing agitator shafts?

Single mechanical seal

1) has a rotating stirrer shaft & rotating ring

2) in the housing is a fixed static ring -> while rotating these 2 parts sliding on each other & spring presses sliding ring against static ring


principle: 2 rings slide on each other (friction & heat creation)

= is NOT 100% sealed = not 100% safe: contamination get through sealing point

-> cannot be used for aseptic purposes & hazardous media -> used for non-sterile/ non-hazardous product/ pumps

 -> microbial contamination can get through sliding point


(6) Sterility testing

(a) Explain the effect of air pockets within the sterilization load during steam sterilization.


(b) How is the pressure in an autoclave related to the amount of noncondensable gases in the autoclave chamber?


(c) How is venting / air elimination performed in a steam sterilization process that uses the gravity or flow principle?


(a) Explain the effect of air pockets within the sterilization load during steam sterilization.

-> when saturated steam is coming into an autoclave = it reaches a cold product surface & gets condensate because the steam cools quickly down & transfers all its saved energy to the product -> so product heats up

-> condensation = very fast & powerful energy transfer process

-> air prevent steam for reaching the product -> bad sterilization


b) How is the pressure in an autoclave related to the amount of noncondensable gases in the autoclave chamber?

limit < 40ml/kg non-condensable gases

3) Pressure increase: In a closed system the pressure rises up (because of high energy & steam needs more space) so the vaporization curve reaching 212°C and it becomes a saturated steam

-> the lower the pressure, the lower amount of non-condensable gases

= if pressure rises to high, then there is a lot if air left causing overpressure


(c) How is venting / air elimination performed in a steam sterilization process that uses the gravity or flow principle?

4) Elimination of air -> by gravity/ flow principle

 - In the beginning product is cold & contains a lot of air

 - air must be eliminated otherwise steam cannot reach product; no condensation

 - air is 1,5 times heavier than the steam = moves down & get pressed out of the system

 - it reaches air separator (Z) with a valve -> at 100°C no cold air flows out & valve closes

5) more steam generation -> pressure inside rises to achieve p = 2 bar & 121,1 °C

(11) Sterile filtration (2 points)

(a) When may sterile filtration be used to sterilize a liquid medicinal product?


(b) As part of the validation of a sterile filtration process the bacterial challengetest (BCT) must be performed. How does it work?

Note: the viability test doesnot need to be addressed.

(a) When may sterile filtration be used to sterilize a liquid medicinal product?

= if product is not heat stable at all, use sterile filtration

= if steam sterilization is not possible then sterile filtration with microbial retentive filter is the last option

=> must be combined with aseptic processing

 with presterilized containers -> only final sterile filtration & filling is done in class A



(b) As part of the validation of a sterile filtration process the bacterial challengetest (BCT) must be performed. How does it work?

Note: the viability test doesnot need to be addressed.


Bacterial Challenge Test, BCT – biological test for filtration

1) Preparation: performance of Viability test (with reference microorganism Brevundimonas diminuta)

= checking if product has bactericidal effect -> germ killing 1log level in 45mins

-> if product is bactericidal = then it needs to get filtered first & then with adding germ suspension

-> if product isn’t bactericidal = germs get added generating a solution with > 107cfu/cm² filter area

2) Provide filtration process = under worst case conditions (with product, filter elements, contact time, pressure, temperature)-> check the filter retention capacity = validation of retention of germs on membrane filters

3) Checking sterility after filtration = 100 % retention of germs (acceptance criterion) -> no single cell of reference germ should pass the filter!

(10) Gamma irradiation

Gamma irradiation and gassing with ethylene oxide are often used for the sterilisation of entire pallets with consumables.

(a) What is the mechanism of action of gamma irradiation?


(b) Give a reason why gamma irradiation is preferred over electron irradiation for this application.



(c) Name one advantage and one disadvantage of gamma irradiation compared to sterilization with ethylene oxide.


(d) Give a reason why ethylene oxide is used for that application but not hydrogen peroxide.(2.5 points)

(a) What is the mechanism of action of gamma irradiation?

= use of high-energy gamma rays to disrupt the DNA or RNA of microorganisms, making them incapable of reproduction

- Generation of Ionizing Radiation: = cobalt-60 creates during radioactive decay high-energy gamma rays

- Penetration of Materials = Gamma rays have high energy and short wavelengths, allowing them to penetrate deeply into materials (sterilization)

- Ionization of Atoms & Molecules = As gamma rays pass through a material, causing ionization of atoms -> results formation of highly reactive ions & free radicals -> killing germs


(b) Give a reason why gamma irradiation is preferred over electron irradiation for this application.

electron irradiation = Limited penetration, but sterilization of products in final packaging possible

Gamma irradiation = High penetration capability sterilization of products in final packaging -> higher assurance level


(c) Name one advantage and one disadvantage of gamma irradiation compared to sterilization with ethylene oxide.


Gamma Irradiation

  • High penetration capability sterilization of products in final packaging & no residuces

  • Problem of radio active waste

Advantages - ETO

1) Broad spectrum of activity

2) Alkylation of nucleic acids, proteins (requires H2O)

3) Excellent penetration capacity


Disadvantages - ETO

1) Carcinogenic

2) Extremely flammable, forms explosive mixtures with air

3) Leaves toxic residues -> not allowed in food sector


(d) Give a reason why ethylene oxide is used for that application but not hydrogen peroxide = sterilisation of entire pallets with consumables.

because: ETO has excellent penetration capacity & broad spetrum of activitity -> better for entire pallets with undefined consumables


hydrogen peroxide

= Very strong oxidizing agent

= Highly corrosive

= Penetration capacity limited






(14) In preparation for aseptic filling of a medicinal product, the formulated product is usually first subjected to germ-reducing filtration and then finally to sterile filtration in class A.

(a) Why does the formulation of the product have to be carried out at least in clean room class C, when subsequently two filtration steps are carried out to achieve sterility?

(b) What must be checked on each filter before and after use?

(1.5 points)

(a) Why does the formulation of the product have to be carried out at least in clean room class C, when subsequently two filtration steps are carried out to achieve sterility?

= to ensure a low & controlled bioburden before getting to the filtration

-> according to EMA-guidline = Frequent bioburden limit: ≤ 10 CFU / 100 ml -> to fulfill this clean room conditions necessary

= the lower the bioburden, the slower cake formation during filtration (during dead-end filtration)

(b) What must be checked on each filter before and after use?

= Filter integrity testing– GMP Guidline

= The integrity of the filter must be checked before & after use

-> after usage = easy to perform -> filter can get taken out for a quality control in lab or in production area

-> before usage = is critical because it must be assured that filter doesn’t get contaminated -> 2 possibilities to keep filter sterile:

1) sterilization in autoclave after filter testing -> validation important

2) filter gets installed in production area & gets sterilized in place -> filter testing can be done in place

(18) Preparation of equipment and primary packaging material

(2 points + 0.5 bonus points for part d)

(a) Why is primary packaging material considered more critical then secondary packaging material?

(b) Name one way in which two silicone tubes can be connected aseptically.

(c) Why are cartridges siliconized on the inside?

(d) At which point in the process does siliconization take place?

(a) Why is primary packaging material considered more critical then secondary packaging material?

= its the first layer of packaging that is in direct contact with pharmaceutical product (e.g. vials/ ampoules/ syringes/ stoppers/ caps)

-> shouldn’t cause recontamination or have negative impact on product -> must be sterile (depyrogenation) / inert/ abrasive proof etc.

(b) Name one way in which two silicone tubes can be connected aseptically.

Disposal systems: Aseptic connection

Aseptic connection via

1) Sterile connectors/ couplings = presterilized system with membraned (gets taken out) for product transfer

2) Silicone hose connectors = used in class A environment of 2 sterile tubes

3) Sterile welding of thermos plastic tubes = 2 tubes get brought together & closed in a special machine (welder) -> inside machine: tubes get cut & welded to each other (+200°C) -> class A air supply

(c) Why are cartridges siliconized on the inside?

= siliconisation with PDMS = polydimethylsiloxane

• very important for Syringes & cartridges

= get covered with anti-friction coating (silicone) to minimize break loose & gliding forces

  • without siliconization stopper doesn’t move at all due to high friction or stopper suddenly moves & stops -> syringe doesn’t work well, no correct dosage (insulin) -> dangerous for patient

break loose force = force that is needed to start movement of a stoper

gliding force = force that is needed to keep stopper moving

(d) At which point in the process does siliconization take place?

1) siliconisation of packaging (syringe/ cartridge) usually before sterilisation (after washing & drying) -> often applied backed on siliconization which has formation of silicon layer during (hot air) sterilization

2) siliconisation of syringe & cartridge stoppers = optionally after sterilization -> stoppers must move to achieve equal distribution of silicone

(12) Sterile filtration (2.5 points)


(b) In preparation for the aseptic filling of a liquid medicinal product, the drug is usually subjected to two filtration steps. The second step is sterile filtration. What is the purpose of the first filtration step?


(c) Sterile filters must be examined for leachables. What are leachables?

(b) In preparation for the aseptic filling of a liquid medicinal product, the drug is usually subjected to two filtration steps. The second step is sterile filtration. What is the purpose of the first filtration step?


bioburden reduction prefilters and/or sterilising grade filters may be used at multiple points during manufacturing process to ensure a low & controlled bioburden before getting to the final sterilising filter

Purposes:

- to prevent the build up of filter cake, first filter takes the main germs load out of product

- if the first filter is having a lack the second filter enables the filtration (safety of process)

- increases the process safety to ensure a sterile product

- separation of the 2 steps is possible, so one is done in class C area & the other in the final filtration


(c) Sterile filters must be examined for leachables. What are leachables?

Leachables (if required) = only required if toxic extractables detected

= Substances leached from the filter material in contact with the product under standard process conditions

= substances that can migrate or leach from materials used in the manufacturing process into the final product

-> for sterile filters, leachables refer to any chemicals or substances that may be released from the filter material and contaminate the filtered product.

 



(12) Validation of sterilization processes (2 points)

(a) How do you call the release of terminally sterilized products without batch-related sterility testing, based solely on the measurement of physical process parameters?


(b) What is checked during the OQ of a steam sterilizer?


(c) What is the acceptance criterion for the Bacterial Challenge Test (BCT) that must be performed as part of the validation of a sterile filtration process?


(a) How do you call the release of terminally sterilized products without batch-related sterility testing, based solely on the measurement of physical process parameters?

Parametric release – Ph.Eur.5.1.1

= can be applied when for the terminal sterilization of pharmaceuticals steam sterilization, dry heat sterilization or ionising radiation are used

-> Precondition: sterilization processes must be validated & be reliable = to get permission by authorities

 

What is parametric release?

= Release of terminally sterilized products without batch-related sterility testing -> for results of sterility testing 14 days needed

= release of product is based solely on the measurement of physical process parameters -> monitoring of parameters

= possible because standard conditions are defined via regulations & are performed in reality with overkilling phase (far above minimal requirements, reaching low SAL) which guarantees sterilisation safety

-> standard conditions (121 °C / 15 min steam sterilization) represent overkill = killing efficiency  sterilization safety

-> But must be approved by the authorities


b) What is checked during the OQ of a steam sterilizer?

OQ = Operational Qualification

  = Checking basic functions of equipment within the working area including chamber integrity (pressure & vacuum test)


(c) What is the acceptance criterion for the Bacterial Challenge Test (BCT)?

Checking sterility after filtration = 100 % retention of germs (acceptance criterion) -> no single cell of reference germ should pass the filter!


(17) Preparation of equipment and primary packaging material (2 points)

(a) Which medium is used in a CIP process for the final rinse of a format part that is later used in an aseptic process?


(b) Name one way in which two silicone tubes can be aseptically connected.


(c) What process takes place for cartridges after washing before treatment in the hot air sterilization tunnel?


(d) What is the final process step of an automated stopper treatment?

(a) Which medium is used in a CIP process for the final rinse of a format part that is later used in an aseptic process?

= is a Validated automatic cleaning without disassembly

->Mobile or fixed systems with connection to SIP

 

CIP – steps = 2 water qualities required

1) Pre-rinse with purified water (pre-rinse)

2) If necessary, wash cycle with (alkaline) detergent

3) If necessary, rinse with purified water

4) Rinsing with water for injection (final rinse)

5) Drying -> otherwise can cause dilution of product



(b) Name one way in which two silicone tubes can be aseptically connected.

= aseptic connection via:

Sterile welding of thermos plastic tubes = 2 tubes get brought together & closed in a special machine (welder) -> inside machine: tubes get cut & welded to each other -> class A air supply


(c) What process takes place for cartridges after washing before treatment in the hot air sterilization tunnel?

= siliconization


(d) What is the final process step of an automated stopper treatment?

Capping/ Crimping = after placing a stopper on a vial, capping must be perform to fully seal the vial

 = ONLY after this step container is considered ready & sealed -> can leave class A

(20) Barrier systems (2.5 points)

(a) RABS and isolators are operated by glove/sleeve systems. What checks/tests must be performed on the gloves before usage?

(b) Describe the principle of a transfer into an isolator during operation using a decontamination airlock.

(c) Indicate one way in which a continuous transfer of material can be made out of an isolator?

a) What checks/tests must be performed on the gloves before usage?

1) Visual inspection of the gloves -> if leak detected no further test needed

2) Leak testing:

 2.1) Pressure decay test = gloves get blowed up with pressure

 2.2) airflow test at constant pressure = air gets inside of gloves & measurement of volume that is needed to keep pressure constant -> if tight, no additional air needed


b) principle of a transfer into an isolator during operation using a decontamination airlock.

3.1) Decontamination airlock/ transfer hatch

= has 2 decontamination airlocks (small chambers) on each side of the isolator that are connected to working area

-> material can get inside & out of the isolator

-> double door system

airlock = Luftschleuse


c) Indicate one way in which a continuous transfer of material can be made out of an isolator?

Open isolator systems = allows continuous material flow, hermetic seal for decontamination, with unidirectional air flow

 = designed for continuous or semi-continuous ingress and/or egress of materials (Ein-/ Austritt) during operations through one or more openings -> mouse holes


Mousehole – transfer system

= for continuous material transfer between different pressure zones within the isolator or out of isolator

-> Pressure tight closure for decontamination

-> fit exactly to the maximum dimension of the transported material (e.g. vials)

  • Active mousehole = with suction (additional protection) Ò overpressure cascade inside isolator




(18) Preparation of equipment and primary packaging material (2.5 points)

(a) In CIP processes, it is important that no spray shadows occur. What is meant by this?


(b) Stoppers that will later be used in syringes in an aseptic process are to be prepared. Outline the process steps of automatic stopper treatment (excluding transfer and storage processes).


(c) What process takes place for syringes after final rinsing and blow-out with compressed air before treatment in the hot air sterilization tunnel?

(a) In CIP processes, it is important that no spray shadows occur. What is meant by this?

"spray shadows" = areas or surfaces within a processing system that are not effectively reached or cleaned by the spray nozzles during the CIP cycle.


(b) Stoppers that will later be used in syringes in an aseptic process are to be prepared. Outline the process steps of automatic stopper treatment (excluding transfer and storage processes).

Automatic stopper treatment

1) Washing & cleaning = to remove any contaminants, particles, or residues.

2) Rinsing = After cleaning, the stoppers undergo rinsing to eliminate any remaining cleaning agents or impurities

3) Drying = to remove moisture

4) Sterilization = autoclaving, gamma irradiation

5) Siliconization = application of a thin layer of silicone oil to the stopper surfaces. This is done to facilitate the movement of the stoppers within the vial or syrange and reduce frction & enebale correct dosing

6) Inspection and Quality Control = to ensure that quality: visual inspection for defects, dimensional checks, and testing for sterility.

7) insert stopper in syringe (with pipe or vacuum)



(c) What process takes place for syringes after final rinsing and blow-out with compressed air before treatment in the hot air sterilization tunnel? = siliconization

(18) Aseptic filling (2.5 points)

(a) Name two dosing systems besides the peristaltic pump that are used for aseptic filling.


(b) What is the purpose of a stopper sorting bowl? Name two aspects.


(c) What needs to be considered when placing a lyophilization stopper following aseptic filling?

(a) Name two dosing systems besides the peristaltic pump that are used for aseptic filling.

1) Rotary piston pump = Drehkolbenpumpe: for temperature sensitive product

= used for: viscous products & for high precision filling (low volumes)

-> not for: metal sensitive products & quick product changes

 Principle:

- cylinder (no valve) & a piston that moves up/ down

- groove of upper part can rotate

- product get inside, groove rotates (180°C) & get filled to final container

 

2) Rolling diaphragm pump = multi-use or single use pumps

- intake port open & product gets inside, piston moves inside

- discharge port opens & product gets out, piston moves out

- diaphragm inside housing covers completely the piston,

no connection to outside (closed system)


(b) What is the purpose of a stopper sorting bowl? Name two aspects

Feeding of stoppers (Zufuhr von Stopfen) = Singularization (Vereinzelung) & correct stopper alignment with (vibratory) sorting bowl

- inside barrier system get transported to a vibratory sorting bowl (is rotation & vibrating)

- only stoppers that are located in right direct/ position are get transferred for capping


(c) What needs to be considered when placing a lyophilization stopper following aseptic filling?

Lyophilization = performed in free dryer under class A conditions

-> because: - vials have special lyophilization stopper (with opening), so the solvent/ vapor can get out

  - opening in stopper = risk of contamination -> class A surrounding & CIP/SIP capable (steam sterilization/ H2O2)

(23) Media fill (3 points)

(a) In which documents is the execution of a media fill specified?


(b) How often / with what frequency must a media fill be repeated after successful initial validation?


(c) An SOP for aseptic formulation and filling of a product requires that the product is freeze-dried and nitrogen is applied for inertisation before closing the stopper after freeze-drying. How do you take these process steps into account for a media fill?

(a) In which documents is the execution of a media fill specified?

= Validation plan


(b) How often / with what frequency must a media fill be repeated after successful initial validation?

Frequency - EU GMP Annex 1

-> APS should be performed as initial validation = including 3 successful APS that cover all working shifts when significant modification in process occurs = again initial validation is required

-> periodic re-validation = APS should be repeated twice a year (2x/ year) per shift & process

-> Every employee who is routinely involved in aseptic processes must participate in 1x successful APS min. 1x/ year with an appropriate job profile.


c) An SOP for aseptic formulation and filling of a product requires that the product is freeze-dried and nitrogen is applied for inertisation before closing the stopper after freeze-drying. How do you take these process steps into account for a media fill?


Lyophilizates -> partial drying process

- FDA recommends that unsealed containers be exposed to partial evacuation of the chamber that simulates the process.

- Vials should not be frozen, and all precautions should be taken so that microbial growth get supported (for evaluation)

(21) Preparation of equipment and primary packaging material (2.5 points)


(b) Which test is performed in the pharmaceutical sector for the validation of CIP processes?


(c) When in the course of the process does siliconization of glass syringes take place?


(d) Name one disadvantage that oily siliconization has compared to baked-on siliconization.


(e) For which pharmaceutical products is the final drying of the stoppers after automatic stopper treatment of particular importance?

(b) Which test is performed in the pharmaceutical sector for the validation of CIP processes?

Validation = Riboflavin test (at least 2 times): flush is fluorescence (green) substance & good water soluble

1) contamination of the inner surfaces with a riboflavin solution

2) drying

3) CIP verification of successful cleaning with UV lamp

-> no green fluorescence should be visible! otherwise dirt left


(c) When in the course of the process does siliconization of glass syringes take place?

= before sterilization -> baked-on siliconization preferred & to reduce friction while transport/ sterilization


d) Name one disadvantage that oily siliconization has compared to baked-on siliconization.

oily siliconization = thick layer -> may cause mechanicall problems e.g. in syrange dur to limited space

-> backed-on silionization = due to treatment by heat (during sterilization) mire robust silocon layer results, lower risk of silicone migration


e) For which pharmaceutical products is the final drying of the stoppers after automatic stopper treatment of particular importance?

Lyophilized (Freeze-Dried) Products = are highly sensitive to moisture

= Dual-chamber technology - syringes

• Chamber 1 = lyophilized drug

• Chamber 2 = diluent for reconstitution

-> Used because: dilution of drug is easier,

safer & higher dosage precision

(15) Validation of sterilization processes (1.5 points)

(a) List four different non-economic requirements that should be included in a user requirement specification (URS) for an autoclave.


(b) What must be considered in the PQ of a steam sterilization process with regard to loading - in addition to the loading schemes of routine production?

(a) List four different non-economic requirements that should be included in a user requirement specification (URS) for an autoclave


1) Underlying regulatory requirements for the system

·   Standards to be complied with, guidelines

2) Location / Environment / Dimensioning

·   Place of installation/ Environment / zone requirements

·   Constructional conditions, e.g. available area, floor load

·   Dimensions (chamber volume)

3) Process

·    Procedure / operating mode (e.g. FRVV)

·    Special requirements for the process (e.g. accelerated cooling)

·    Design / Type, (cylindrical/rectangular; horizontal/vertical)

·    Working range (pressure / temperature)

·   Performance / energy consumption / energy efficiency

4) Monitoring / Measurement technology

·    Sensors/ Accuracy/  Additional features



(b) What must be considered in the PQ of a steam sterilization process with regard to loading - in addition to the loading schemes of routine production?


5) PQ = Performance Qualification = Verification of sterilization success based on relevant loads -> actual process validation

= provide evidence that equipment works according to specified criteria (reliable sterilization)

= include physical & biological tests

 

Requirements for PQ (Verification of sterilization process)

Inspections must verify that:

1) the documentation confirms a successful IQ & OQ -> IQ/ OQ & cycle development should be successfully finished & working

2) the test load of the sterilizer contains products which are routinely handled -> loading patterns must be defined and [...] represent product families which present the greatest difficulty for the sterilisation -> include worst case conditions

3) the packaging system must be identical to the packaging used for routine production

4) the preconditioning = storage before sterilization must correspond to criteria

5) the loading configuration & amount of load has to bedefined -> size and/or the mass of the sterilizer load


(12) Validation of sterilization processes (2.5 points)


(b) What should be considered when validating a steam sterilisation process with regard to loading?


(c) What is the general name of test strips with bacterial spores used to validate sterilization procedures?

b) What should be considered when validating a steam sterilisation process with regard to loading?

= worst case conditions -> coltest spot



Cycle development = based on relevant loads -> Optimization of steam penetration, venting etc. for different loads

= establishment of sterilization cycles for routine manufacturing for different loads


PQ = Performance Qualification = Verification of sterilization success based on relevant loads -> actual process validation

= provide evidence that equipment works according to specified criteria (reliable sterilization)

= include physical & biological tests

 

Requirements for PQ (Verification of sterilization process)

Inspections must verify that:

1) the documentation confirms a successful IQ & OQ -> IQ/ OQ & cycle development should be successfully finished & working

2) the test load of the sterilizer contains products which are routinely handled -> loading patterns must be defined

  and [...] represent product families which present the greatest difficulty for the sterilisation -> include worst case conditions

3) the packaging system must be identical to the packaging used for routine production

4) the preconditioning = storage before sterilization must correspond to criteria

5) the loading configuration & amount of load has to bedefined -> size and/or the mass of the sterilizer load


c) What is the general name of test strips with bacterial spores used to validate sterilization procedures?

The general name for test strips with bacterial spores used to validate sterilization procedures is "biological indicators" or "biological indicator strips”

These strips contain a known number of highly resistant bacterial spores, such as Bacillus subtilis or Geobacillus stearothermophilus. Biological indicators are employed to assess the efficacy of sterilization processes, for steam (autoclaving), ethylene oxide gas, hydrogen peroxide gas et.


(12) Validation of sterilization processes (2.5 points)

(a) During the DQ you have to select a steam sterilization process for the sterilization of 2 m long silicone tubes used for aseptic filling. Which process do you choose? Give a justification for your choice.


(b) What is the name of the document in which the user demands on a sterilization system are recorded as part of the DQ?


(c) What is checked during thermoelectric testing with an empty sterilization chamber as part of the PQ of a steam sterilization process?

(a) During the DQ you have to select a steam sterilization process for the sterilization of 2 m long silicone tubes used for aseptic filling. Which process do you choose? Give a justification for your choice.


DQ = Design Qualification = Determination of requirements for the equipment ->(URS = user requirements specifications & functional specification document)

Ionizing radiation sterilization -> E-beam tunnel

= its applicabel for tube & E-beam tunnel is long, has enough space for 2 m logn tube


(b) What is the name of the document in which the user demands on a sterilization system are recorded as part of the DQ?

URS = user requirements specifications & functional specification document


(c) What is checked during thermoelectric testing with an empty sterilization chamber as part of the PQ of a steam sterilization process?

1) Thermo-electric test with empty chamber

= temperature and pressure sensors get distributed inside empty chamber to check uniformity of parameters (°C/ ∆p)

 -> minimal deviations (+/- 2°C; +/-10 kPa) allowed -> even coldest spot must fulfil criteria

 -> must be performed for minimum & maximum cycle times and temperatures

(14)

A silicone tubing shall be used for the aseptic filling of a drug. The preparation of the silicone tubing consists of the following steps:

[1] In class D it is placed in a double-door washing machine, washed therein, and then removed to class C.

[2] In class C the hoses are double-wrapped in tyvek bags and loaded into a pass-through autoclave.

[3] The removal of the sterilized hoses takes place in class B.

[4] The hoses are then transferred to a previously decontaminated RABS. During transfer, the door of the RABS is briefly opened and the outer tyvek bag is removed.

[5] Finally, with the doors closed, glove ports are used to open the inner tyvek bag and install the tubing.

How do you evaluate the process?

Are all steps (1 to 5) in compliance with the requirements of annex 1 and a minimal contamination risk for the drug processed in class A?

Is there potential for optimization? If yes, which?

(3 points)

How do you evaluate the process?

= the process doesnt fulfill the criteria of aseptic preparation

Are all steps (1 to 5) in compliance with the requirements of annex 1 and a minimal contamination risk for the drug processed in class A?

= No are not, high contamination risk through wrong unpacking in wrong zone

Is there potential for optimization? If yes, which?

[1] ist right

[2] & [3] = tubs get unpacked/ opened (removal of second packaging tyvek bag) in class C -> surface sterilization from C to B is correct

[4] door of RABS shouldn't get opened -> high contamination risk (class A inside) & unpackeging of second packaging layer has to be in class C

[5] doors shouldn't open (transporttation via decontamiantion airlock or RTP better) - inner tyvek bag better to open in class B area & transfer sterile tub to class A for final filling/ processing


Main steps:

I) Class C/D: 1) tubs get mechanically unpacked/ opened -> HEPA filtered air

 2) tubs get introduced into E-beam tunnel -> through airlock doors

 3) tubs get transferred to irradiation zone -> surface sterilisation

II) Class B: 4) tubs get transported trough class B zone to class A

III) Class A: 5) filling zone -> presterilized syringes/ vials get filled