1. Describe the basic principle of chromatography.
Stationary phase, mobile phase, different polarities
Analytes (components of the mixture) have different affinities (depends on size, charge, etc. physical-chemical properties) to the stationary phase and therefore different retention times —> separation of analytes.
2. Explain the terms “analyte” and “eluent”!
Analyte: compound(s) of interest, substance mixture
Eluent: mobile phase (liquid/gas)
3. Give one example of how metabolites can be derivatized for GC-MS. Name two specific advantages.
Whole lipids cannot be analyzed by GC/MS because they are often not volatile + bigger than 600 kDa
Transmethylation (cleaving fatty acid from head group of the lipid and esterifying with methyl group instead): enhances stability and thermostability of the compound, helps detection (increased sensitivity), increases volatility.
DMOX: helps determining location of double bonds (promotes equal fragmentation of fatty acids, DBs are located where the difference between two peaks is 12 instead of 14).
Trimethylsinylation: important to make polar compounds suitable for GC-MS (decreases polarity, increases volatility).
4. Which materials represent the mobile and the stationary phase by gas chromatography (GC)?
Specify two advantages and one disadvantage of GC.
Mobile phase: inert carrier gas (N2, Ar, He)
Stationary phase: long thin column coated with silica gel
Advantages: retention times are reproducible across labs (when working with a column of the same size and coating), no packing problems as in LC, high resolution.
Disadvantage: compounds must be thermostable, mostly unpolar and smaller than 600 kDa.
For non-targeted approaches, LC is more suitable because it can handle compounds bigger than 600 kDa.
One of the main advantages of GC is the retention time stability.
5. Name two types of detectors, that are usually coupled with GC.
FID (flame ionization detection)
MS (mass spectrometry)
6. How does a flame ionization detector work? What are the advantages and disadvantages.
After separation, a compound is burned in a hydrogen flame, leading to the formation of ions. They are detected by electrodes and the signal is amplified. (The resulting current is proportional to the number of C atoms.) To characterize a compound, you need a standard to compare your results to. FID does not provide structural information like MS or NMR, but is much more sensitive and suitable for quantification (FID: accurate over seven orders of magnitude, MS: only 5 orders of magnitude).
7. Explain the abbreviation “HPLC”. What is the difference to other chromatographic methods?
High-Performance Liquid Chromatography
HPLC uses high backpressure to pump analytes/mobile phase through column, the column is packed with silica (in normal-phase HPLC). HPLC has very high chromatographic resolution and uses particles of sizes around 3.5µm (UHPLC: 1.7µm, even more precise signal).
“Shorter”/”steeper” gradient of the mobile phase leads to shorter retention times.
Tight, precise packing leads to higher separation.
Difference: PRESSURE (compared to GC and other LCs), liquid mobile phase (compared to GC), suitable for a broader range of compounds (compared to GC which has restrictions in size and polarity), lower temperature needed (than GC).
8. What is the characteristic feature of reversed-phase HPLC?
Normal: unmodified silica (polar) as stationary phase, solvent gradient from unpolar to polar.
Reverse-phase: unpolar stationary phase (e.g., silica esterified with acyl groups), mobile phase starts with polar solvent.
9. Which types of analytes elute first by separation of a complex metabolite extract by normal phase-HPLC and which one by reversed phase-HPLC?
Normal phase: most unpolar
Reversed phase: most polar
10. Mention three differences concerning the mobile and the stationary phase and the gradient between gas chromatography and liquid chromatography.
Mobile phase: inert gas for GC, solvents for LC
Stationary phase: GC columns are coated with silica, LC columns are packed with silica
Gradient: GC uses a temperature gradient, LC a solvent gradient for elution.
11. Name the three general parts of a mass spectrometer and describe the main function of each part.
(Very impotant question)
Ion source: creates detectable ions (in the gas phase!) from formerly uncharged molecules, potentially fragmentation (depending on ionization technique)
Mass analyzer: separation of ions based on their m/z value.
Detector: detection of ions and their m/z value (duh)
12. Describe two principles of producing ions for mass spectrometric analyses. Explain the basic principles and the chromatographic methods they can be combined with.
EI and ESI:
EI is combined with GC
- EI produces M+ radicals
ESI is combined with LC (you need a liquid to produce the ions)
- ESI produces mostly adducts à complicated analysis for unknown compounds
- ESI can be used in positive or negative mode, ions are either [M+H]+, [M+Na]+,… or [M-H+]-, [M+formate]-, [M+acetate]- …
- Some compounds can be ionized in both modes, some only in one – therefore, it is important to do both modes to detect everything (in non-targeted approach)
13. What is the major difference between a mass spectrum obtained by electron impact (EI) ionization and by electrospray ionization (ESI)?
EI (electron ray shot at molecule) causes fragmentation, ESI does not. ESI produces adducts, EI does not.
14. Define the terms: total ion chromatogram/count, extracted ion chromatogram/count and mass spectrum.
(IMPORTANT)
Total ion chromatogram: all counts for all time points for all m/z
Extracted ion chromatogram: counts for all time points, but only for one m/z
Mass spectrum: all m/z for one time point
Possible exam question: Several mass spectra of different time points are given, TIC and EIC must be developed from the mass spectra (addition of counts per minute for each peak and time point for TIC, addition of counts per minute for one m/z for each time point).
15. How does tandem mass spectrometry based on a triple quadrupole work? Explain three modes of analysis! Which information could be obtained with the different modes?
A QqQ consists of two mass analyzers with a collision cell in between. Both mass analyzers can operate in either scan or select mode; a scan provides separation of all analytes according to their m/z, select results in the selection of a compound of one specific m/z. In the collision cell, the compound(s) is/are fragmented.
Depending on how scan/select modes of the two mass analyzers are combined, different kinds of analyses can be done:
1) The m/z of a compound of interest is known but its structure is unknown. Product ion scan can be used to deduce the structure (e.g. which fatty acids are part of a specific lipid?). It can also be used to generate tables for 4).
2) Precursor ion scan can be used to find compounds with a specific precursor in different kinds of metabolites, for example all lipids containing PC. In Q1, the different metabolites are separated, Q2 leads to fragmentation of the metabolites (in this example, leading to PC fragments), Q3 detects the fragments of interest (PC). Since Q1 and Q3 are interconnected, it is possible to find out which m/z values resulting from Q1 exhibit a PC peak in Q3.
3) Similar research question as in 2) (?), but different approach: The fragment of interest is indicative for a specific lipid class (e.g., PC). To obtain knowledge about how many precursors (e.g., PCs) are in the mixture, Q1 is in scan mode. In Q3, the loss of a specific fragment is searched for, not the fragment itself. Afterwards, you can deduce which of the scanned compounds in Q1 showed the specific fragment you are interested in.
4) Scenario: You have a mixture of lipids and want to find out what is in it. You need a reference table with masses of the lipids you want to detect (!). For every lipid, you run the SRM, selecting for a certain m/z and confirming the identity with the select run. You are “blind” for all the lipids that are not in your table. Anyway, this is a very fast and sensitive method.
16. Specify the benefits and limitations of an ESI-TOF-MS analysis.
ESI-TOF-MS principle:
ESI is a soft ionization technique that does not cause fragmentation.
TOF (time of flight) is a separation technique that involves a long tube in which the ions are separated by the time they require to pass through (the longer the tube the better the resolution).
MS: mass analysis
Benefits:
High resolution and sensitivity
High mass accuracy
Determines wide mass range
Limitations:
Complex samples: difficult to identify individual components
Expensive
Detector saturation for too high concentrations
What is it?
Total ion chromatogram
What are these?
Extracted ion chromatograms