Explain how a gas-filled detector works.

Depending on supplied voltage -> 3 different modes:

  1. Ionization chamber (lowest voltage)

    The voltage on ionization chambers is just sufficient to separate the positive (ions) and negative charges (electrons) produced in the consequence of the interaction of the radiation particles with the detector gas and to avoid their recombination.

    Used for accurate dosimetry measurements (e.g. for calibrating radiation fields) and for the measurement of pulsed radiation

  2. Proportional counters

    The voltage applied to a proportional counter is of that magnitude that the produced charged particles get accelerated towards the anode and the cathode respectively. On their way through the counting gas of the tube, the primary charged particles are able to ionize further gas atoms or molecules which results in an amplification of the separated charge

    Therefore proportional counters are well suited for the discrimination between alpha- and beta particles for example in contamination monitors.

  3. Geiger-Müller tube (highest supply voltage) (most common):

    Independent from energy of the penetrating radiation particle a charge avalanche is generated in the tube because of the high voltage conditions. Therefore amplification factors in the range of 106 to 108 are possible


    ! Dead time of the GM-counter: This dead time usually is in the range of 10 to 100 µs and describes the time period the detector is not able to process another radiation event after a successful avalanche generation.